mixed Balmer and calibration module wavelength calibration

Forums Spectroscopy DSLR spectra Processing in ISIS mixed Balmer and calibration module wavelength calibration

#579070
Robin Leadbeater
Participant

Note that the calibration assistant uses both calibration lamp lines and Balmer lines, which is explained in more detail in method 3 of the above link, so both sets of lines have to be clear in the spectra and identified correctly by ISIS. Checking the individual errors for each line can help with finding which lines are causing the problem.

Cheers

Robin