Robin LeadbeaterParticipant
I wondered this too. A piece of Mg alloy not large enough to have been tracked but enough to produce a trail ?
Robin
Robin LeadbeaterParticipantMichael Richmond’s on line calculator can be used to estimate the performance of a given setup for example
http://spiff.rit.edu/richmond/signal.shtml
Robin
Robin LeadbeaterParticipantadded replies to #9 and #11 above
(Note to webmaster – we need a quote button so we can keep to a linear thread and not lose replies back up the thread)
Robin
Robin LeadbeaterParticipantThen you are currently slightly undersampled already and would not want to go any shorter in focal length. Undersampling in photometry is much less desirable than oversampling.
Robin
Robin LeadbeaterParticipantYep sorry, factor of 4 out but the ratio (and hence the increased counts) is the same. Changing the magnification (ie focal length) changes the counts per pixel (both star and sky background counts) but makes no different to the counts in the aperture, either the star counts or the sky background counts. The only difference is if you spread the light over more pixelsthan you need to, you increase the camera noise contribution. In practise this is only the thermal noise, not the read noise as you can compensate for this (in CCD cameras) by binning pixels. Thermal noise is very low in modern CCD, particularly at typical exposure times used for photometry so the net effect is that changing focal length does not improve photometry SNR, only increasing the aperture does this
Robin
Robin LeadbeaterParticipantWhat is your typical star size with the 70mm f6 currently? 3 arcsec is only 6.1um with your existing setup so you are potentially undersampled already so there would be no advantage going to a lower focal length to beat the seeing. (I think potential gain is only marginal anyway as all you do by moving to a shorter focal length is reduce the camera noise contribution, the star and sky background counts stay the same).
The extra aperture will gain you some photons though (Area of C6 is 608 cm2 allowing for the central obstruction compared with 154 cm2 so ~4x or ~1.5 magnitudes)
Robin
Robin LeadbeaterParticipantWould the C6 with a focal reducer to f6.3 perhaps be a better bet? (3 arcsec at 950mm focal length = 13.8um)
Robin
Robin LeadbeaterParticipant3 arcsec at 290mm focal length is just 4.2um so with most cameras you are likely to be undersampled unless you defocus aren’t you ?
Robin
Robin LeadbeaterParticipantHi Bill,
>I think this is one of the reasons that the technique of comparing particular line ratio’s (as devised by Borovicka) is about as good as we can get.
Even this needs a relative flux calibration though so instrument response including flat field issues and extinction still need to be considered.
My only foray into this branch of spectroscopy was back in 2005 and I was glad to at least get something before moving on to other targets
http://www.threehillsobservatory.co.uk/astro/spectra_20.htm
My mentor back then was Ed Majden who had been doing meteor spectroscopy as far back as the ’70s at least, using film, then video. Do you know if Is he still with us ? His website is down and the AMS website mentions him doing work in the 50’s so he must be getting on
https://www.amsmeteors.org/ams-programs/meteor-spectroscopy/meteor-studies-at-majden-observatory/
Your continuous monitoring though is certainly taking things to a new level, building statistically useful numbers of observations and working outside the normal showers.
It is an area where new developments in technology are moving in our favour too, with affordable large CMOS chips with low read noise, fast readout and improved bit depth compared with the old 8 bit cameras. High efficiency volume phase transmission gratings is another interesting development, though I believe they are still rather specialist/expensive.
Cheers
Robin
Robin LeadbeaterParticipantHi Bill,
There is a link with her email at the bottom of the download page
http://astrosurf.com/vdesnoux/download.html
Cheers
Robin
Robin LeadbeaterParticipantFlux calibrating these spectra even in relative flux is not trivial. Yes the standard method to correct for instrument response and extinction using a reference star measurement is ok in principle but there are a few extra things to watch out for which make it tricky when using this technique with wide field moving targets like this. Specifically flat fielding, background subtraction and differential extinction.
Flat field correction of any slitless spectra is complex as each point in the field is a combination of zero orders and diffracted light from other points in the field so there is no one to one correspondence like in conventional flats. In practise this is effectively impossible to untangle. With static targets you can get round this problem by measuring the reference and target spectra in the same position in the field, which is the way I recommend using the Star Analyser for example but this is obviously not possible for meteors of course so my suggestion would be to take a series of spectra of a bright standard star at different locations in the field and asses exactly how much effect it has on the spectrum. If the spectrum shape varies significantly then some allowance has to be made for this depending on the location of the meteor spectrum.
Similarly sky background subtraction is difficult compared with narrow field fixed targets where the sky can be measured directly above and below the spectrum. Perhaps this is not too much of a problem for short exposure videos of meteors though where subtraction of frames before and after the meteor could be used. (Linearity of the light response is obviously important – no gamma correction to be used)
Extinction effects over such a wide field can be significant and will vary along the trail. These could be corrected for using an atmospheric model and some sort of mean elevation figure for the meteor though.
Robin
Robin LeadbeaterParticipantHi Bill
Just checked the Visual Spec website and all looks fine this end. Was it a particular link on the tutorial page? (Some of them use flash which perhaps could have been flagged up as a security problem. ) If there is a problem then we need to let Valerie know.
Cheers
Robin
Robin LeadbeaterParticipantI quickly pulled out a couple of instrument responses at H alpha (using a 1200 l/mm grating so covering a wider wavelength range than the 2400 grating.) They are flat within +-1.5%
Robin LeadbeaterParticipantYou could try measuring the IR using a high reference resolution spectrum and see how it compares with you MILES IR.(Unless I am working low to the horizon or at the blue end just use a bright star like Vega, Altair, Regulus for example which can be recorded quickly.) If it is significantly different, it might be safer to just rectify the spectra you have already taken. (You can still submit them the BeSS setting the appropriate flag in the fits header.
To be a bit controversial (and this is just my personal view) I think for relative flux calibrated H alpha spectra, most of the time IR correction of a narrow wavelength range at high resolution is a waste of good observing time. (And may even lead to increased variability, though I need to quantify this). If you use a flat, You are already getting rid of all instrument affects as they divide out, leaving the flat lamp spectrum and the atmospheric extinction which hardly varies across the range. If you are using ISIS, this even removes the small slope due to flat lamp spectrum, assuming a black body at 2750K I believe so the IR ends up being effectively a horizontal flat line. In the projects I have been involved in where narrow range spectra have been used, the first step in analysing the data has been to rectify all the spectra first in any case.
Cheers
Robin
Robin LeadbeaterParticipantIt is worth noting that the technique of using a reference star near in elevation to the target is technique developed to make it easier for amateurs. Because professionals have stable setups and know their atmospheric conditions, they tend to use a standard instrument response measured infrequently using precisely measured spectrophotometric standard stars, combined with a measure of extinction on the night together with an atmospheric model. ISIS does have the tools to do this though.
http://www.astrosurf.com/buil/isis/guide_response/method.htm
Robin LeadbeaterParticipantHi Kevin,
You can do an instrument response in the usual way but you need a high resolution spectrum to compare with so you can match profile including the line. You can see an example of that about halfway down on Christian Buil’s page on reducing LHIRES spectra here
http://www.astrosurf.com/buil/isis/guide_lhires/tuto1_en.htm
and also specifically covered on this page
http://www.astrosurf.com/aras/tutorial5/note1_us.htm
Fortunately unless you are working at the far blue end, atmospheric extinction does not have much effect over the narrow wavelength range so you can use stars which might be some distance from the target (or even at a pinch instrument responses taken on different nights)
Two good sources for high resolution spectra of bright stars are the UVES bright stars
http://www.eso.org/sci/observing/tools/uvespop/bright_stars_uptonow.html
and provided you are working above 4000A,the ELODIE 3.1 list (spectra from the ELODIE archive selected for quality.) See here on ARAS for more background on this set of stars
http://www.spectro-aras.com/forum/viewtopic.php?f=8&t=1369
Both of these are available in the built in ISIS database (you have to load the ELODIE 3.1 star list)
http://www.spectro-aras.com/forum/viewtopic.php?f=8&t=1864
Is it necessary/worth it rather than just rectifying the spectrum for small wavelength ranges? Possibly not, it depends on the application but BeSS recommends it.
Cheers
Robin
Robin LeadbeaterParticipantYep the good old Relco 480 (but possibly gold plated given the price !) To be fair though, I believe there is a lot of variability so Shelyak select the better ones and discard the rest. Starlight Xpress use it in their spectrograph too. A great discovery by Richard Walker who should be on commission! My first tests with it in the LHIRES are here
http://www.spectro-aras.com/forum/viewtopic.php?f=8&t=606&hilit=relco+starter#p2380
Robin
Robin LeadbeaterParticipant> I know that early versions had three degrees of freedom on the guide mirror (two tilt and back-and-forth). The reduction to one with the piston design has thrown some of the ‘baby out with the bathwater’ in my view..
I am thinking that too. I can see why it was done as It was easy to get into a mess with the old tip/tilt setup but the new setup depends on that wall of the spectrograph being very accurately square which cannot be guaranteed with the type of case construction used. Having reviewed what I posted on that ARAS thread, I am not convinced my alignment is quite as good as originally. There is not enough room for the old mechanism now though with the the calibration lamp mechanism.
Robin LeadbeaterParticipantAndy’s comments started me wondering about this. There are two ways to move the spectrum vertical position in the camera field. You can adjust the spectrograph mirror or you can move the star along the slit. You commented, Kevin that the slit position and the sweet spot did not correspond to the centre of the guider field. (They should do when everything is correct.) I wonder if because of this, where you are placing the star is off the spectrograph axis, exaggerating the widening effect in the spectrum.
Robin
Robin LeadbeaterParticipantI have converted over to the latest guider mirror setup and I made some tests today using the internal flat lamp and a pinhole in place of the slit (actually a piece of foil with a slit in it overlaid at right angles over the middle of the original slit.) This effectively simulates an on axis star in the spectrograph, though the effective focal ratio may not be correct.
I first checked that the crossed slits were central in the guider image so the slit and my guider image should be centralised on axis. (Hopefully the guider image coma will be minimum at this position too when I make a test on the sky)
I then recorded flat and calibration lamp spectra at H alpha 1/3 from top,middle and 1/3 from bottom of the camera field (having previously focused the calibration lamp lines for best focus.)
The first thing to note is the instruction to set the spectrum at 1/3 (from top or bottom) is rather meaningless as the actual position will depend on the size of the sensor. (I have an ATIK 314 with 6.45um pixels so moving 1/3 of the field equates to moving ~2.2mm)
I found that there was no change in total flux or spectrum resolution for all three positions. (though because the effective focal ratio of the flat lamp may not match that of the scope, potential vignetting will need to be checked using a real star)
I did indeed find that the thickness of the spectrum increases when moving the spectrum from top1/3 to bottom1/3. In this case approximately doubling in width (FWHM) from ~ 50um to ~100um or 8-16 pixels (The actual size of the pinhole in this direction is no known as it was just scored in the foil) This seems less than you were quoting Kevin, though perhaps you have a larger sensor/smaller pixels ?
I also made the same measurements on the zero order image with similar results, though the height of the zero order image was a bit less than that of the spectrum, (~2/3) possibly due to the anamorphic factor which makes the diffracted image narrower in the dispersion direction but wider at right angles to it.
Robin
-