Robin Leadbeater

Forum Replies Created

Viewing 20 posts - 701 through 720 (of 1,123 total)
  • Author
    Posts
  • in reply to: What wavelength error is acceptable with an Alpy? #580178
    Robin Leadbeater
    Participant

    Here are my errors using the ISIS automatic calibration on lamp lines only (Heliocentric and RV corrected). It confirms what I have generally been finding with low errors at the red end increasing at the blue end.  As expected the lines in the lamp spectrum are in exactly the right place  (to better than +-0.5A).  I also had a look at Hugh’s lines in my calibrated lamp spectrum and where different lines are used they are still exactly in the right place so it does not seem to be dependent on the lines used. (Note I could not use the 6562.85 lamp  line (H alpha) as the Hydrogen has all leaked from my lamp which is a few years old now)

    Cheers

    Robin

    lamp lines HD206165 ALPY calib
    wavelength error
    7383.98 7065.2 -0.1
    7147.04 6678.2 0.2
    6506.53 6562.8 -0.3
    6266.49 5875.7 0.2
    5852.49 4921.9 -0.5
    5400.56 4861.3 -0.7
    4965.08 4387.9 -0.6
    4764.87 4340.5 -0.5
    4657.9 4101.7 -1.0
    4545.05 4026.2 -2.3
    4510.73 3970.1 -3.0
    4158.59 3888.6 -2.8
    3946.1 3835.4 -3.6
    3797.9 -4.0
    3770.6 -4.0
    3750.1 -4.2
    in reply to: What wavelength error is acceptable with an Alpy? #580176
    Robin Leadbeater
    Participant

    Hi David

    “Including the 3946.097A line gives a considerably worse fit at the blue end.” 

    I have had a look at the high resolution echelle spectrum of the lamp that Shelyak use in Richard Walker’s atlas. (plate 98 order 57)  The 3946.097 Ar line that ISIS uses for a lamp line fit with the ALPY is in fact a close triplet 3946.097, 3947.505, 3948.979  so definitely incorrect.  I don’t think this is connected with what John is seeing but I suspect this may be the cause of my increasing error towards the blue using just the lamp calibration.

    I took a spectrum of MILES star HD206165 last night to experiment with. It is a B2i star so lots of narrow H and He lines which hopefully should show up any errors

    Cheers

    Robin

     

    in reply to: What wavelength error is acceptable with an Alpy? #580139
    Robin Leadbeater
    Participant

    A good point.  Buil seems to have done a reasonable job of selecting evenly spaced lines for the fit just using the lamp though and the fit residuals are much smaller than the errors seen. The use of a 4th order fit though does make me slightly nervous that we might be “fitting elephants” .

    lamp lines used  (EDIT 2018-11-02    Note these are the fit values, not the published wavelengths )

    3946.331
    4158.342
    4510.673
    4544.769
    4657.878
    4765.023
    4965.325
    5400.742
    5852.471
    6266.287
    6506.282
    7147.661
    7383.626
    in reply to: What wavelength error is acceptable with an Alpy? #580166
    Robin Leadbeater
    Participant

    Hi David,

    The ALPY calibration module has always had the NeAr lamp. The link I posted was specifically for calibration the ALPY. The first line used in the  ALPY calibration is 3946.1 and the calibration deteriorates below there. eg

     http://www.astrosurf.com/buil/isis/guide_alpy/calibration/fig_fr22.png

    What  is the first line you use with the LISA? (note some lines which might be resolved at the LISA resolution could be blends with the ALPY so not useable)

    When calibrating the ALPY with the NeAr lamp, ISIS definitely reports the fit values not the published values. You can see examples on the page linked. For example

    http://www.astrosurf.com/buil/isis/guide_alpy/calibration/fig_fr19.png

    and for a bad calibration

    http://www.astrosurf.com/buil/isis/guide_alpy/calibration/fig_fr23.png

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580167
    Robin Leadbeater
    Participant

    I like Andrew’s idea of calibrating using the lamp and then ofsetting the spectrum to match the star lines. This is similar to what is done with the LHIRES where telluric lines at H alpha are used to correct for any offset.  You would have to consider the RV of the star though and for highest accuracy perhaps the Heliocentric correction.

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580164
    Robin Leadbeater
    Participant

    Yes I only use the lamp lines. The fit does deteriorate below 3900A though. (The lowest line ISIS uses) which is why Buil added the mixed star and lamp option

    http://www.astrosurf.com/buil/isis/guide_alpy/resume_calibration.htm

    see comparison of methods 2 and 3

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580158
    Robin Leadbeater
    Participant

    Hi John,

    I wouldn’t hold back from submitting spectra. (You could have a look in the databases for other ALPY spectra to see how their calibration looks.) I think a wavelength calibration accuracy of ~1/4 the resolution is still good (At the end of the day my ALPY spectra using just the lamp lines are no more accurate. The errors are just at the other end of the wavelength range). There is still something niggling to do with the difference in how the ALPY sees the lamp compared with the star to nail down here but I suspect it will turn out to be something common to ALPYs in general.

    Dont refocus to sharpen up the lines in the far red to the detriment of the blue end. The ALPY performance is biased towards the blue end where it actually outperforms the LISA

    I think ISIS is finding the lines OK so exposing more probably will not make much difference but exposing for the brightest line that ISIS uses is a good idea and if you want to try going further you can sum several exposures to increase the signal/noise in the faint lines in the same way as summing flats works  (I used to do this but the results using just one well exposed lamp exposure seemed to work just as well.)

    Cheers

    Robin 

    in reply to: What wavelength error is acceptable with an Alpy? #580155
    Robin Leadbeater
    Participant

    Hi John

    Last night’s specrum of Pi Andromeda (HD 00369) was still about 3 A out at H alpha but very accurate from H beta to epsilon. “

    I presume this was using the mix of lamp lines and Balmer lines for calibration?.     If so, the Balmer lines from H Beta down will always be in exactly right the right place because ISIS used them for calibration. 

    Similarly if you just used the lamp lines for calibration and produce a spectrum of the lamp, I suspect you will find that the lamp lines are in exactly the right place (within the accuracy of the fit ~+-0.5A)  

    The problem arises when we use lamp lines to calibrate star spectra.  There is a shift somewhere between the star spectrum and lamp spectrum.  Based on the latest result this does not seem to be a stability problem.

    There are differences between the optical paths for lamp and star which could produce small systematic shifts. We may have found the limit of absolute accuracy of using the internal lamp for calibration. (note this is a systematic error, not a lack of precision/repeatability  so not a problem when studying changes in wavelength eg in radial velocity measurements of binary stars.) 

    To take this further I would say we would first need to compare results from other ALPYs to determine if it is a limitation of the instrument or if this particular example is worse for some reason. Since I am seeing signs of similar size errors in star lines with my ALPY when using just the internal lamp, I suspect it could be a limit in absolute accuracy. In which case there is no need to be overly concerned but longer term there may be some modifications that could be made to the ALPY or the calibration technique to improve the calibration accuracy further.  

    Using a mix of lamp and star for calibration confuses the issue. I would say The definitive test would be for ALPY owners to take a spectrum of a star with known reliable RV, Calibrate just using the internal lamp and look for errors in the star lines.

    This could be a good project for the workshop

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580156
    Robin Leadbeater
    Participant

    Hi John,

     “I’ve processed the calibration image as is if were a star, of course I had to use the same image to calibrate it. The result was odd, but the neon 6965 line was incorrectly calibrated.”

    The spectrum plot looks odd. Did you tick “sky not removed”  If you don’t do this it subtracts the lines from them selves, leaving almost nothing behind)

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580149
    Robin Leadbeater
    Participant

    Hi David,

    Ah sorry, I see the wavelengths I posted (and listed by ISIS in the reports) are the values calculated from the fit, not the true values. 

    There does not seem to be a problem with the quality of the fit to the lines, at least within the range of lines used. (RMS ~0.2-0.3A seems typical for the ALPY.)  The issue is apparent systematic differences in wavelength up to 3A between the lamp  and the star)

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580145
    Robin Leadbeater
    Participant

    Hi Andrew,

    The extrapolation is into the red not the blue ?

    Do you have the residuals for the blue points for the different fits? Particularly how much improvement in the fit is there going from 3 to  4 ?

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580136
    Robin Leadbeater
    Participant

    Ok the reflector position looks ok. ( I was concerned that the reflector might not be rotating to the correct position. It can be adjusted by operating  the switches in a particular sequence) Not sure what any offset in the lamp might do (The asymmetry in the lamp internally is normal) and I cant see mine at the moment as it is mounted on the scope but is probably best not to change anything until you have checked the calibration with  lamp spectra taken at the same time as the star spectrum.

    I have been having a look at some of my past calibrations and I am also seeing  wavelength errors between lamp and star (I  use just the lamp without the star Balmer lines) In my case H alpha is spot on but there is an increasing error towards the blue  to around 2-3A by 3900A where the lamp lines end, so errors of this order might be typical. What are other other ALPY users seeing?

    We might be seeing a fundamental limitation in the calibration accuracy using the internal lamp.  At this level (~1/4 the slit width) effects such as the position of the star on the slit and subtle differences in line shape due to optical aberrations come into play.

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580132
    Robin Leadbeater
    Participant

    One other thing that could be worth checking is that the servo controlled reflector is set so the calibration lamp is shining directly down on the slit. If the slit is illuminated at a significant angle, it can cause small shifts relative to the star spectrum. (I’ve not heard of  this with the ALPY but I suspected this problem with the original manual LHIRES system)

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580131
    Robin Leadbeater
    Participant

    OK,   I think the constant shift with just the lamp confirms that there was some shift in the spectrograph (It is very small. I think 3A equates to about 5um or about a pixel).  The changing temperature might have been a factor. With the ALPY I  usually take a lamp spectrum each time I change target  during an observing run even though I normally dont see any shift during the night. (With the LHIRES I  see a lot of movement so I take a lamp spectrum at start and end on every target and every half hour or so during long exposures but it is much less stable thermally and mechanically than the ALPY)

    Cheers

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580125
    Robin Leadbeater
    Participant

    OK John,

    ISIS seems to have found the lines ok and the fit looks good. I am wondering if there has been some sort of shift  between the lamp spectrum and the star spectrum.  With the mixed Balmer and lamp line calibration you are using, the lines used for calibration up to 4861 (H Beta) are the Balmer lines in the star and the lines higher than that are from the calibration lamp. Normally the ALPY is very stable but if everything is not quite tightened up, there could have been a shift between the star and lamp exposures. Did you take the lamp spectrum under the same conditions as the star ? 

    Can you also run the calibration using just the lamp lines ? If there was a shift, we should then see all the Balmer lines shifted, not just H alpha

    Cheers

    Robin 

    in reply to: What wavelength error is acceptable with an Alpy? #580123
    Robin Leadbeater
    Participant

    OK I have checked the MILES zet Dra spectrum and the Balmer lines do appear to be at the expected wavelengths ( eg 4861, 6563A) There will be small adjustments to make for the radial velocity and heliocentric correction but these will not add up to as much as 3A so there looks to be something going on here.  Can you post a copy of what ISIS reports in the “Go” window please?

    Thanks

    Robin

    in reply to: What wavelength error is acceptable with an Alpy? #580122
    Robin Leadbeater
    Participant

    Hi John,

    I am not sure the wavelength calibration of the MILES stars is all that accurate. I remember seeing some apparent discrepancies when I was trying to do some radial velocity work.  Do your individual errors for the lamp lines look good particularly those bracketing the H alpha line? (from the report that ISIS generates in the “Go” window. )   

    I would not worry about adjusting the Grism further if ISIS is finding the lines to do the wavelength calibration OK (The RMS error should be low, <0.5A and there should be no obvious larger errors in any individual line as reported by ISIS)

    The kink in the instrument response of the ALPY around 4000A is real and often seen, though the cause is not clear. See my recent post here for example.

    https://britastro.org/comment/5542#comment-5542

    Cheers

    Robin

    in reply to: BAA Practical Spectroscopy Workshops #580121
    Robin Leadbeater
    Participant

    In this case though what we need is the geometric median which minimises the distance travelled.  My first student vacation job was with the GPO, writing a computer program to decide where the international exchange should be sited to minimise cabling costs (distance from exchanges, weighted by the traffic). I think it came out near Reading (This was in the 60’s and the computer CPU time was charged by the second) 

    in reply to: BAA Practical Spectroscopy Workshops #580120
    Robin Leadbeater
    Participant

    Kevin does have a valid claim to the centre of population though. 

    https://en.wikipedia.org/wiki/Appleby_Parva

    in reply to: BASS queries and Andy’s Tutorial. #580113
    Robin Leadbeater
    Participant

    I filter the spectra to match the resolution and shift them slightly if needed to line them up before division. With MILES spectra the match is generally very good then and little editing is needed (I edit out any artifacts rather than choosing points on the curve as that way I can smooth less and include more subtle effects in the instrument response.  eg as here)

    https://britastro.org/comment/5542#comment-5542

    Cheers

    Robin

Viewing 20 posts - 701 through 720 (of 1,123 total)